Problem 8- Calibration and Sensitivity (10 points) The following data was taken
ID: 3369264 • Letter: P
Question
Problem 8- Calibration and Sensitivity (10 points) The following data was taken from a pressure transducer - Pressure Output (Pa) 0.5 0.4 10 20 50 100 2.3 6.9 15.8 36.4 110.1 253.2 It has been plotted below- Pressure Transducer Calibration Curve 300 250 200 150 5 100 50 y 0.007x2+ 1.8556x-1.8307 80 100 120 02060 50 Pressure (Pa) Using the polynomial equation, estimate the sensitivity of the pressure transducer at a pressure of 50 Pa. a. Compare this sensitivity with that obtained by a linear fit through the endpoints of the data set. b.Explanation / Answer
Solutiona;:
poynomial eq is
y=0.007x^2+1.8556x-1.8307
for x=50
y=0.007*50^2+1.8556*50-1.8307
y=108.4493
voltage =108.4493 volts for pressure =50 Pa
Solutionb:
sensitivity =110.1 V for pressure=50 Pa
predicted sensitivity=108.4493 V for pressure =50 Pa
Residual=predicted-observed
=108.4493-110.1
=-1.6507
A negative rsidual means the models prediction is too large(the model over estimates) the sensitivity for pressure=50 pa