An analytical experiment yields a calibration curve which has a linear least squ
ID: 554781 • Letter: A
Question
An analytical experiment yields a calibration curve which has a linear least squares fit of
S = 1.25CPb + 0.06
with
S = relative instrument signal
CPb = lead concentration (in ppb)
When a blank sample (CPb = 0.0 ppb) was analyzed 25 times, the mean analytical signal was found to be 0.12 with a standard deviation of 0.019. Based on the above calibration equation and experimental results:
a) What is the calibration sensitivity?
b) What is the signal limit of detection for this method?
c) What is the limit of detection in concentration units for this method?
d) If the signal measured for a particular sample is S= 205+17, what is the lead concentration?
Please do include the formulas
Explanation / Answer
a) Calibration Sensitivity = Slope of the Calibration Curve
Calibration Sensitivity = 1.25
b) Signal limit of detection for this method = 3 x Stadard Deviation
Signal limit of detection for this method = 3 x 0.019 = 0.057
c) Limit of detection in concentration units for this method = 10 x 0.019 = 0.19
d) Lead Concentration = 205 / 1.25 = 160 ppb