Newborn infants are screened for cystic fibrosis by measuring the chloride in th

ID: 889333 • Letter: N

Question

Newborn infants are screened for cystic fibrosis by measuring the chloride in their sweat. Pilocarpine, a sweat inducer, is applied to a small spot on the skin. The sweat is than collected and analyzed using an ion selective electrode. If the chloride is above 70 mMol/liter, there is a strong suspicion of cystic fibrosis. The schematic of the one piece electrode is below:

         Ag/AgCl reference// Cl- selective cell

Note the selective membrane is the cathode!

The potential when the electrode is placed in a 0.065 Mol/liter solution is 0.2235 volt. The potential of the electrode in the infant's sweat is 0.2329 volt. What is the chloride ion concentration?

Explanation / Answer

Since it is cathode,

reaction occuring is:

1/2 Cl2 + e-   --->Cl-

E= Eo - 0.059 * log [Cl-]

case 1:

Potantial of 0.065 Mol/liter solution is 0.2235 volt

0.2235= Eo - 0.059 * log [0.065] .....eqn 1

case 2:

Potantial of 0.2329 volt in infant. Let concentration be M

0.2329 = Eo - 0.059 * log [M] ....eqn 2

subtract eqn 1 from eqn 2

0.2329 - 0.2235 = Eo - 0.059 * log [M] - { Eo - 0.059 * log [0.065] }

0.0094 = 0.059 * log [0.065] - 0.059 * log [M]

0.0094 = (-0.07004) - 0.059 * log [M]

0.059 * log [M] = -0.07944

M=0.045 M

Answer:0.045 M